Henry Dore, Elizabeth Rendon-Morales, Rodrigo Aviles-Espinosa, FPGA Implementation of ECG Signal Processing for use in a Neonatal Heart Rate Monitoring System, in Proceedings of 9th International Electronic Conference on Sensors and Applications, 1 November–15 November 2022, MDPI: Basel, Switzerland, doi: 10.3390/ecsa-9-13258
Share
Email
Facebook
Twitter
LinkedIn

FPGA Implementation of ECG Signal Processing for use in a Neonatal Heart Rate Monitoring System

image
1. University of Sussex, School of Engineering and Informatics, Robotics and Mechatronics Systems Research Group, UK
2. Senior Lecturer in Electrical and Electronic Engineering (Engineering and Design), University of Sussex, Falmer, UK
3. Deputy Head of Department of Engineering, School of Engineering and Informatics, University of Sussex, Falmer, UK
4. Senior Lecturer (Engineering and Design) School of Engineering and Informatics, University of Sussex, Falmer, UK
Abstract

An FPGA based hardware accelerator for bio-signal digital filtering in a neonatal heart rate monitoring system employing electric potential sensors (EPS) is presented. These active sensors provide a non-contact alternative to traditional ECG electrodes, but are more susceptible to noise such as power line interference and motion artefacts, therefore additional filtering capacity is required.

The proposed system contains a single hardware filter stage for antialiasing, with the remaining digital signal processing required to provide a clinical standard ECG performed on an FPGA (National Instruments myRIO 1900). This is compared with a previous microprocessor version (Raspberry Pi 3, BCM2837 processor) containing a dual hardware/software filtering scheme, with the aim of simplifying the analog front end and allowing for reconfigurable filtering in the digital domain. A custom neonate phantom was employed to emulate real world conditions and ambient noise.

The developed FPGA system was shown to have a signal quality comparable with the microprocessor implementation, with an average signal to noise ratio loss of 2%. A 12 dB increase in attenuation of the predominant 50 Hz noise and a 90% reduction in energy per sample filtered was shown compared to the microprocessor version, indicating both efficiency and filter effectiveness gains. The proposed system accurately calculated the heart rate of a simulated neonatal ECG signal, with lower heart rate variation than the microprocessor system. Finally, the phantom was used to broadcast data from the preterm infant cardio-respiratory signals database (PICSDB) and the FPGA filtering scheme was shown to remove the majority of the ambient 50 Hz noise with an average reduction of 30 db, and provide a clean ECG signal.

These results demonstrate that FPGA filtered EPS ECGs have comparable signal quality to the combined HW/SW filtering implementation, with a reduction in complexity and power consumption.

Keywords
ECG
electric potential sensor
FPGA
filtering
noise
power line interference
neonatal ECG