Field-Effect Transistor-Based Biosensor for pH Sensing and IgG Detection

image
1. Institute of Biophysics and Biomedical Engineering, Faculty of Sciences, University of Lisbon, Portugal
2. Exotictarget, Viana do Castelo, Portugal
3. Instituto de Engenharia de Sistemas e Computadores - Microsistemas e Nanotecnologias, Lisbon, Portugal
4. Exotictarget, Viana do Castelo, Portugal, Portugal
5. Instituto de Engenharia de Sistemas e Computadores - Microsistemas e Nanotecnologias, Lisbon, Portugal, Portugal
Abstract

Introduction: Maintaining the balance of urine pH and urine metabolite concentrations is vital for a good health, with abnormal levels indicating various conditions such as infection, kidney disease, or metabolic disorders. Hence, this work aims to create a biosensor using field-effect transistors (Bio-FET) to measure urine pH and detect multiple metabolites, starting with Immunoglobulin G (IgG).

Methods: Our Bio-FET system comprises a microfabricated gold electrode, working as an Extended Gate (EG) and a readout circuit for signal detection and amplification. pH tests were conducted with six distinct solutions (10.34, 8.60, 7.40, 7.00, 4.00, and 2.91) applied onto the EG surface, resulting in Ids – Vref transfer curves performed within the linear region of a MOSFET, allowing us to create a calibration curve (Ids – pH). For the IgG detection, an Anti-IgG was covalently bound to a mixed 11-Mercaptoundecanoic acid (MUA) and 11-Mercapto-1-undecanol (MUD) (ratio 1:2) Self-Assembled Monolayer (SAM) on the gold electrodes. Each step of the surface functionalization was characterized using ellipsometry. Preliminary tests were conducted to the Bio-FET response when exposed to two distinct solutions: (1) phosphate-buffered saline (PBS) solution at pH 7.40; and (2) PBS solution spiked with IgG (0.01 mg/mL).

Results: Regarding pH tests, the results showed a linear relationship, wherein elevated pH values were associated with increased current values, with a corresponding sensitivity of 2.20 µA/pH. The binding of target IgG with the specific antibodies immobilized on the EG surface triggered a positive vertical shift of the PBS-IgG solution transfer curve when compared to the one obtained when the PBS solution alone was used.

Conclusions: The Bio-FET system shows a good response both for pH measurement and IgG detection. However, regarding IgG detection, further tests using different PBS-IgG solutions are required to evaluate the limit-of-detection and linear range of the Bio-FET system as an IgG sensor.

Keywords
Bio-FET system
pH tests
Extended-Gate
MOSFET
Biofunctionalization
IgG
Oral Presentation