A dual-mode vertical flow assay for simultaneous bacteria identification and load assessment in urine

Shu-Yun Sheu 1
image
image
1. Institute of Biomedical Engineering, National Tsing Hua University, Hsinchu 300
2. Department of Pediatrics, National Cheng Kung University Hospital, College of Medicine, National Cheng Kung University, Tainan 701
Abstract

Rapid urine-based bacteria screening requires not only pathogen identification but also estimation of total bacteria burden. However, most paper-based assays provide only one of these outputs. Here, we present a dual-mode vertical flow assay (VFA) designed to simultaneously perform species-specific bacteria identification and total bacteria load assessment from a single urine sample.
The platform integrates two complementary colorimetric modules within one vertical flow device: an HRP–antibody/TMB immunodetection module for species-specific recognition and a WST-8–mPMS metabolic module for quantifying viable bacteria load. These reactions are arranged in spatially separated detection spots to minimize chemical and optical interference while enabling parallel readouts under a simple vertical-flow workflow. A graded porous membrane stack was used to regulate capillary-driven transport, improve residence time at the detection interface, and support consistent color development.
Using Escherichia coli as the target bacteria, the immunodetection module produced concentration-dependent signals in both tryptic soy broth and human urine, achieving stronger quantitative fitting in urine with Hill and linear regression coefficients of R2 = 0.8643 and R2 = 0.8764, respectively. The WST-8–mPMS metabolic module further enabled total bacteria load assessment across different bacterial species, including E. coli and Staphylococcus aureus, with linear fitting coefficients ranging from R2 = 0.72 to 0.84 across broth and urine conditions. In the integrated device, both modules generated concentration-dependent responses from a single urine sample, demonstrating threshold-oriented detection over 5 × 104 to 108 CFU/mL and covering the clinically relevant bacteriuria level of 105 CFU/mL.
These results demonstrate that dual-mode VFA can provide complementary information on bacteria identity and overall bacteria burden within a single assay format. This strategy offers a practical framework for rapid urine bacteria screening and may support future development of integrated paper-based diagnostic platforms.

Keywords
Vertical flow assay
Bacteria detection
Species-specific identification
Total bacterial load
Urine screening
Paper-based diagnostics