EventsThe 2nd International Conference on Bioengineering
Published
This submission belongs to the session S5. Biomechanics and Sports Medicine of the event The 2nd International Conference on Bioengineering
Published date
16 Nov, 2026
Contribution&Funding
Methodology,
[Anne Lizbeth Carrillo Chuyes] ; [Hector Alonso Garcia Paz]
Software,
[Anne Lizbeth Carrillo Chuyes] ; [Hector Alonso Garcia Paz]
Validation,
[Anne Lizbeth Carrillo Chuyes] ; [Hector Alonso Garcia Paz]
Investigation,
[Anne Lizbeth Carrillo Chuyes]
Funding: --
Citation
Anne Lizbeth Carrillo Chuyes, Hector Alonso Garcia Paz, Low-Cost Baropodometric Platform Based on Velostat for Instrumented Evaluation of the Modified Star Excursion Balance Test, in Proceedings of The 2nd International Conference on Bioengineering, Barcelona, 10 November–13 November 2026, MDPI: Basel, Switzerland
Share
Email
Facebook
Twitter
LinkedIn

Low-Cost Baropodometric Platform Based on Velostat for Instrumented Evaluation of the Modified Star Excursion Balance Test

Hector Alonso Garcia Paz 1
1. Biomedical Engineering – Faculty of Engineering, Peruvian University of Applied Sciences (UPC), Lima, 15023, Peru
Abstract

The Modified Star Excursion Balance Test (mSEBT) evaluates dynamic postural control in athletes, but its conventional scoring depends only on reach distance and gives no objective information on plantar pressure distribution or on the displacement of the center of pressure (CoP) of the support foot. Commercial force and pressure platforms can provide these variables but are usually expensive, rely on proprietary software, and have limited portability outside the laboratory, restricting accessibility in sports and academic settings. This work aimed to design and develop a low-cost baropodometric prototype, based on a Velostat piezoresistive sensor array, integrating plantar pressure detection, CoP estimation, and mSEBT reach-distance measurement in a single platform.

The prototype, measuring 40 × 40 cm, employs a 15×15 Velostat piezoresistive array, 225 sensing points, with copper strips forming rows and columns, scanned via two shift registers and two analog multiplexers, and digitized by the ESP32's ADC. Three ultrasonic sensors measure reach distance in the anterior, posteromedial, and posterolateral directions of the mSEBT. A custom PCB integrates the acquisition electronics, while the Python software performs baseline correction, filtering, spatial smoothing, and weighted-average CoP calculation to obtain direction-specific trajectory metrics.

Functional tests confirmed that the array distinguishes loaded from unloaded cells and localizes pressure across the 225 cells, validating end-to-end signal acquisition. A real-time graphical interface displays pressure maps and CoP trajectories, with data export. Validation tests, performed by shifting the participant's centre of mass across the platform while standing on both feet, showed continuous mediolateral and anteroposterior CoP displacement; repeating the test with the loading axis rotated 90 degrees relative to the array's row/column wiring yielded equivalent, continuous trajectories, indicating no directionally-dependent crosstalk bias. Residual row/column variability was observed, and CoP values remain preliminary, reflecting relative rather than calibrated absolute pressure, pending per-cell calibration against a reference system.

A functional, integrated low-cost prototype was achieved, capable of simultaneously acquiring plantar pressure distribution, estimating CoP trajectory, and recording mSEBT reach distances. Combining these variables in a single device, the platform is a promising complementary tool for quantitative dynamic-stability assessment in sports biomechanics, pending absolute calibration against reference systems.

Keywords
Center of pressure
plantar pressure
piezoresistive sensor
sports biomechanics
From Clinical Need to Commercialization: Building Translational Pathways for Point-of-Care Diagnostics