S.M. Amin Arefi, Youjie Zhang, Lucas Maignaut, Christian Kastrup, Dana Grecov, Simulation-Driven Development and Bench Testing of a Clog-Free Endoscopic Spray Device for Upper Gastrointestinal Bleeding, in Proceedings of The 2nd International Conference on Bioengineering, Barcelona, 10 November–13 November 2026, MDPI: Basel, Switzerland
Share
Email
Facebook
Twitter
LinkedIn

Simulation-Driven Development and Bench Testing of a Clog-Free Endoscopic Spray Device for Upper Gastrointestinal Bleeding

Youjie Zhang 2
Lucas Maignaut 3
image
1. Department of Mechanical Engineering, University of British Columbia, Vancouver
2. Department of Biomedical Engineering, The Marquette-Medical College of Wisconsin, Milwaukee
3. Department of Mechanical Engineering, National Higher School of Advanced Techniques (ENSTA), Palaiseau
4. Departments of Biochemistry, Biomedical Engineering, and Pharmacology and Toxicology, Medical College of Wisconsin, Milwaukee
5. Department of Surgery, Division of Trauma and Acute Care Surgery, Milwaukee
6. School of Biomedical Engineering, University of British Columbia
Abstract

Introduction: Sever upper gastrointestinal bleeding (UGIB) is a life-threatening condition that often requires rapid and effective endoscopic intervention. Hemostatic powders offer a minimally invasive approach for bleeding control; however, cohesive powders such as CounterFlow can exhibit poor flowability through narrow catheters, increasing the risk of agglomeration, inconsistent dosing, and clogging. This study presents a simulation-guided approach for developing a novel pneumatic spray device for reliable endoscopic delivery of cohesive hemostatic powder.

Methods: The device consists of a funnel-shaped mixing chamber and a centrally positioned reciprocating sieve. Swirling gas flow suspends and disperses the powder, while the sieve regulates powder feeding and disrupts agglomerates before transport through the catheter. Computational fluid dynamics (CFD) simulations were performed in ANSYS Fluent to characterize swirling flow, pressure distribution, and particle transport inside the chamber. In parallel, ANSYS Rocky discrete element method (DEM) simulations were used to evaluate powder sifting behaviour and estimate the powder feeding rate into the swirling flow. Simulation results informed the selection of operating parameters, including gas flow rate and sieve oscillation frequency. Bench testing was then conducted under clinically relevant conditions to assess discharged gas volume, delivered powder mass, spray duration effects, and clog-free operation.

Results: The simulations identified key flow structures responsible for powder suspension and transport and demonstrated how sieve frequency influences powder feeding and dosing uniformity. Bench testing supported the computational findings and showed that the optimized prototype achieved consistent, controllable, and clog-free CounterFlow powder delivery.

Conclusions: CFD and DEM simulations provided critical insight into the coupled effects of swirling gas flow, powder sifting, and dosing control. This simulation-driven framework supports the development of reliable endoscopic spray devices for cohesive hemostatic powder delivery in UGIB treatment.

Keywords
Upper gastrointestinal bleeding
Hemostatic powder delivery
Computational fluid dynamics
Discrete element method
Pneumatic spray device