Introduction: Tuberculosis (TB) remains a leading cause of mortality from infectious diseases worldwide, particularly affecting developing countries such as Peru, Bolivia, and others, making early diagnosis critical to reducing transmission. Traditional Ziehl-Neelsen staining requires complex 1000X magnification and highly trained personnel to identify acid-fast bacilli. In contrast, fluorescent Auramine O staining enhances diagnostic sensitivity, allowing evaluation at 400X–500X and facilitating faster, more accurate identification.
Methods: Overcoming limitations of a previous prototype designed for FITC where a 405 nm laser caused rapid photobleaching due to being far from its absorption peak, a low-cost 3D-printed inverted fluorescence microscope integrated with a smartphone was developed. The system utilizes a 405 nm continuous laser diode at 600 mW output power, aligning closely with the absorption peak of Auramine O to optimize excitation efficiency and achieve an optical spatial resolution of at least 1 µm.
Results and Discussion: An instrumental comparison test was conducted against a commercial Olympus BX41 fluorescence microscope (Olympus Corporation, Tokyo, Japan), equipped with WH10X/22 widefield eyepieces (10x), a 50x objective (500x total magnification), and a fluorescence illumination system, by capturing images of the exact same spot on the fluorophore-stained glass slide. The developed device successfully captured images of Mycobacterium tuberculosis bacilli at 400X with quality comparable to the commercial system, demonstrating improved identification over Ziehl-Neelsen. Due to the proper spectral match with the Auramine O stain, a stable fluorescence signal was maintained for at least 10 seconds under continuous excitation without noticeable photobleaching, mitigating the rapid photobleaching previously observed. The estimated manufacturing cost is approximately 290 USD (excluding the smartphone).
Conclusions: The developed microscope provides a portable, cost-effective, and efficient alternative for TB diagnosis. Furthermore, the use of continuous commercial lasers proves highly advantageous, providing high excitation intensity and precise resolution for resource-limited settings.