This paper proposes a wireless resonant Inductive and Capacitive (LC) sensor for glucose sensing. A sensor composed of a capacitor with interdigital electrodes and an inductor for glucose sensing is presented. The resonance frequency and impedance were measured as the sensing parameters. A glucose beverage concentration from 0 to 44% is used, resulting in a resonance frequency change from 1.9217 MHz to 1.8681 MHz, and the impedance of the sensor changes from 170.33 Ω to 110.68 Ω. The resonance frequency and impedance relationship to glucose beverage concentration are well presented by a decreasing exponential function. Using an exponential regression, the resonance frequency shows an average regression error of 1.38%. Likewise, the impedance shows an average error of 3.47%. The linear range of the sensor is also analyzed at the glucose concentration range between 0 and 4%. The sensor exhibited a sensitivity of 424.6 kHz and 721.6416 Ω, respectively, with a linear regression (r2) of 0.9853 and 0.9553 respectively.
                    Previous Article in event
            
                            Previous Article in session
            
                    
    
                    Next Article in event
            
                            
                                                    
        
                    A Wireless Resonant LC Sensor For Glucose Detection
                
                                    
                
                
                    Published:
01 November 2022
by MDPI
in 9th International Electronic Conference on Sensors and Applications
session Chemo- and Biosensors
                
                                    
                
                
                    Abstract: 
                                    
                        Keywords: Wireless, LC sensor, Interdigital, Glucose sensor
                    
                
                
                
                
        
            